Re: Not capturing anything
Reply #7 –
The OLS is capturing at 200Mhz, which is 5ns resolution. The sample buffer can only hold 120us (0.12ms) of captures total at that rate. At a 50ms SPI signal toggle rate, you won't see anything.
Try a -much- slower sample rate. 32Khz would yield 750msecs of captures.
The trigger causes it to start capturing the instant channel 0 is high. Perhaps try a trigger that looks for a transition instead (ie 0 ->1) would help?
ie: Set stage 1 to detect 0 on ch0, then stage 2 to detect 1 on ch0. Only check the "Action" flag on stage 2 to start captures.
Cheers!
-- IED